Post by WJT-FMC on Jun 19, 2023 17:23:59 GMT
CHECK YOUR WIRES ON BI-BAG BOARD/POWER LOGIC, IF IT IS SLIGHTLY PULLING OR OFF IT WILL CAUSE THIS ERROR
|
|
|
Post by WJT-FMC on Jun 19, 2023 17:25:51 GMT
SORRY..ACUTATOR BOARD
|
|
|
Post by WJT-FMC on Jun 19, 2023 17:26:31 GMT
CHECK YOUR WIRES ON BI-BAG BOARD/ACTUATOR BOARD, IF IT IS SLIGHTLY PULLING OR OFF IT WILL CAUSE THIS ERROR
|
|
|
Post by Fishnqueen56 on Jul 5, 2023 15:23:28 GMT
had the same problem I checked all ribbon cables, replaced brushes in my D-air pump, and calibrated my transducer pressures. that fixed my problem
|
|
|
Post by brocklee on Aug 24, 2023 14:32:03 GMT
usually if the transducer wont calibrate its 106 or 105 when replacing make sure board dont get wet, if no luck from there you move on to valves, next check cable that goes to the interface board before the board its self Thank you for this! I just had this issue myself and went to calibrate the pressure transducer and it gave me an "acid pressure calibration error" or something like that. Switched out V105 and it goes through calibration and rinse! Thank you njftech!!
|
|
|
Post by LuisM on Feb 8, 2024 20:48:09 GMT
its always brushes for me when it says that From Dallas Texas You were right,in my case was brushes on the Dearation motor
|
|