Post by biomedfossil on Aug 29, 2022 20:08:26 GMT
I am a Biomed that works for Fresenius. I am having an issue with chairside on a 3T0S machine. I had to replace the flash card on the iu board. I have tried to downloaded the chairside program from a different machine. The problem machine isn't letting it load up. I have made 3 different flash drives with no luck. It looks like the machine is sensing the flash drive. When I put the flash drive in and turn on the machine. The "upload image" isn't coming up. I wasn't sure if I need to call IT or Fresenius machine support. Any ideas??
|
|
|
Post by Guest on Aug 29, 2022 21:44:01 GMT
I just had this today. It was a loose cable coming from the keyboard. The toggle button would work. I believe it's cable J5.
|
|
|
Post by gnurk on Aug 30, 2022 12:23:56 GMT
thanks for posting your fix
|
|
|
Post by anon's face on Sept 1, 2022 14:53:34 GMT
I am a Biomed that works for Fresenius. I am having an issue with chairside on a 3T0S machine. I had to replace the flash card on the iu board. I have tried to downloaded the chairside program from a different machine. The problem machine isn't letting it load up. I have made 3 different flash drives with no luck. It looks like the machine is sensing the flash drive. When I put the flash drive in and turn on the machine. The "upload image" isn't coming up. I wasn't sure if I need to call IT or Fresenius machine support. Any ideas?? This isn't a loose cable. The images you're making don't have the "boot installer" on them. Fresenius made the ability to grab an image really easy, but left out out the part where the thumb drive needs an installer. You're going to have to contact IT and tell them you need an image LOADED, not copied.
|
|