Post by Jacktripper on Sept 15, 2022 11:49:35 GMT
Yes I have a 2008T mACHINE that has no loading pressure. It pass testes fine but while checking my pressures i realize it has no loading pressure. replaced loading pressure regulator but still no loading pressure? Any ideas
|
|
|
Post by gnurk on Sept 15, 2022 12:19:01 GMT
could be blockage in your hydrochamber
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 231
|
Post by stuff on Sept 15, 2022 13:06:14 GMT
is the gauge working?
|
|
|
Post by gfreely on Sept 15, 2022 13:06:46 GMT
Does it have a Ldg Press when you do a Deaeration cal?
|
|
BMF
Junior Member
Posts: 86
|
Post by BMF on Sept 15, 2022 13:44:51 GMT
As gfreely asks, be sure you have both deaeration and loading pressure when trying to calibration deaeration and loading. If you do but you can't get loading pressure in dialysis, I assume you have a Bibag machine because in order to have a loading pressure reading in dialysis w/bibag, the blue wand needs to be out of it's port to obtain a reading. More info will help everyone give the proper advice. If you can't get a good deaeration reading in calibrate mode, clamp the tubing coming out of V39 at the bottom of the hydrchamber. If you can now obtain good pressure, V39 is either leaking or stuck open.
|
|