Post by Manny on Sept 15, 2022 17:34:17 GMT
Hello I'm on my way to another Clinic to check out a Machine that's pulled for 24v ERROR a few times this Month. The BMT that was there said he checked the power harness, ground wire and reset the boards. anybody have any Recommendations ? maybe need new Board. Ill check back couple days. Thank you !!
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 239
|
Post by stuff on Sept 15, 2022 18:35:22 GMT
When i had machines pulled multiple times for 24v low, 24v high. its was fixed by installing a new motherboard
|
|
|
Post by gfreely on Sept 15, 2022 19:05:23 GMT
I called Big F for a 24V Error a few months back and the guy told me that it is most likely the Pwr Logic Bd, Pwr Supply or Motherboard. We have had the issue twice: once was the board, the other was the pwr supply. Motherboard is the logical go-between...
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Sept 16, 2022 15:09:43 GMT
My best guess would be a power logic board or Motherboard. I had this happen a few months ago, and we discovered that on the motherboard, top left hand side, is the C1+ and C2- capacitor, which was, quite frankly, burnt to hell and back looked it too edited to correct pos and neg
|
|
|
Post by guest1234 on Sept 20, 2022 17:43:39 GMT
I replace at 2 motherboards a month on my machines due to this issue. Now they have never had one prior to this but I do dust my card cage and power supply before replacing but 99.8 percent of the time its your motherboard.
|
|
|
24V ERROR
Feb 23, 2023 1:31:14 GMT
via mobile
Post by G25 on Feb 23, 2023 1:31:14 GMT
Replaced 21 Motherboards on 21 machines with all kinds of voltage issues. Never saw another voltage alarm for 4 years following. I always go Motherboard first.
|
|