Post by planters on Sept 20, 2022 15:14:45 GMT
I have 2008 T machine with 23000 hrs , having problem with conductivity comes up to 12.7 not more than that , calibrated both pumps , checked all pressures everything is good called Fresenius told me to change some boards tried that also , still the same , any clues , all the other parameters are good Temp. TMP all
|
|
eric
Full Member
Posts: 112
|
Post by eric on Sept 20, 2022 18:38:17 GMT
Try a new bicarb pump? if it is a 45 x system, then the acid seems to be coming through and not the bicarb. Maybe a bad diaphragm on the bicarb pump or not fully stroking in dialysis? Is the oring on the bicarb wand or is it worn? using a central feed system? if so, is it plugged with dried bicarb?
|
|
|
Post by dave74 on Sept 20, 2022 19:05:22 GMT
Conductivity at 12.x is a classic sign of leaking or torn balance chamber membranes. I suggest doing the balance chamber membrane leak test. I feel pretty sure the membranes are leaking. Do you know if the membranes have ever been replaced? If all your machines have approximately the same number of hours, you may see a number of membrane failures at the same time.
|
|
|
Post by planters on Sept 20, 2022 19:05:51 GMT
O-rings both wands changed I didn't try new pump , rebuilt both pumps and calibrated , I am using Jugs for Acid and Bicarb
|
|
|
Post by planters on Sept 20, 2022 19:06:58 GMT
B.C Volume is correct
|
|
|
Post by dave74 on Sept 20, 2022 20:11:26 GMT
|
|
|
Post by bimitiri on Sept 21, 2022 2:07:18 GMT
Maybe try the checking or replacing the tubing connecting to the bicarbonate pump and suction wand.
|
|
|
Post by gnurk on Sept 21, 2022 13:47:33 GMT
check to see if your acid and bicarb are made up correctly and not too weak
|
|
|
Post by dave74 on Sept 21, 2022 15:03:29 GMT
True story here. When I took the level 1 class, I asked about the balance chamber membranes. The instructor said not to worry because they last for years and all my machines were new. I have seen the 12.x conductivity many times, and almost every time it has been the membranes. Checking the volume does not check the membranes for leaks.
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Sept 21, 2022 16:10:58 GMT
I agree with dave, sounds like a b/c membrane tear. i just keep an empty bicarb jug i use to put my blue hanson and a small hose for the uf sample port line in, cut the top off an acid jug, clamp off v28's line with a metal hemostat i use specifically for this purpose. Ie: what is says in the TM
|
|
|
Post by planters on Sept 22, 2022 12:00:11 GMT
it was the BC membrane, I changed it works fine , was a real test , The Vol. was exact that why I ruled out the BC , but after you guys advised I changed it , it woks fine , thank you guys for the advise
|
|
|
Post by gnurk on Sept 22, 2022 12:06:33 GMT
to check your bc membranes use the dribble check in the ts guide that is always a good way to tell
|
|
|
Post by Guessed on Sept 22, 2022 12:38:22 GMT
Thanks for posting yer fix
|
|
|
Post by planters on Sept 23, 2022 12:26:46 GMT
Thank You Guys
|
|