Post by jay232 on Sept 22, 2022 15:20:35 GMT
I know this has been brought up but still not fixed. I have a 2008t machine with bibag I am getting a bibag temp sensor error. I have changed out every temp sensor and calibrated. I have swapped known good actuator, piggyback board, and function boards still have issue I ran through full rinse still wont clear. I am thinking the lower dist board on bibag hydro assy but with the FDA hold on parts I can not get one a timely manner. It is a super pain in the ..... to pull and replace the hydro assy or i would swap with one from another machine. Anyone have a suggestion or two?
|
|
|
Post by gnurk on Sept 23, 2022 12:57:23 GMT
call tech services
|
|
|
Post by Bob2 on Oct 31, 2022 12:45:15 GMT
That is not helpful. They read the same book we do. I spent an hour on the phone with tech services, and they had nothing. This particular tech was also field service before. We would not be on here if tech services could fix it.
|
|
BMF
Junior Member
Posts: 86
|
Post by BMF on Oct 31, 2022 13:21:00 GMT
Try a known good NTC3, it is used as a reference for the Bibag temperature.
|
|