Post by Mike Hernandez on Sept 23, 2022 19:22:52 GMT
Good PM , to all of you Guys,
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 148
|
Post by phillipashleyludlow on Sept 23, 2022 19:28:22 GMT
|
|
|
Post by Mike Hernandez on Sept 23, 2022 19:28:50 GMT
Good PM , to all of you Guys, I have a b.braun version 9.12 dialysis machine with this message UF balance air. I replaced the o-rings in the dialysate coupling , but it keeps coming back, and some times the machine will show on the screen Ultrafiltration error check patient weight. any help !!!. will be greatly appreciated. Thank you
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 148
|
Post by phillipashleyludlow on Sept 23, 2022 19:31:32 GMT
Good PM , to all of you Guys, I have a b.braun version 9.12 dialysis machine with this message UF balance air. I replaced the o-rings in the dialysate coupling , but it keeps coming back, and some times the machine will show on the screen Ultrafiltration error check patient weight. any help !!!. will be greatly appreciated. Thank you Admin Seems like your kinda deal
|
|
|
Post by Admin on Sept 24, 2022 11:04:19 GMT
Thanks Phillip! Check the conductivity cells for leakage
|
|
quilla
New Member
Life is Good.
Posts: 25
|
Post by quilla on Oct 19, 2022 16:40:39 GMT
You are right Admin.It has happened to me already. In turned out to be both the ENDLf and BICLF cells cracked in the flow inlets and outlets.
|
|