Post by Guest2 on Sept 26, 2022 17:24:17 GMT
I have a K2 machine that I can't calibrate the temp on. I put in a new heater and my temp never goes above 34 no matter what I do. The odd thing is if I put it in rinse the temp reads 37 and when I heat disinfect it it works fine. Appreciate any help.
|
|
|
Post by robertb on Sept 26, 2022 22:59:44 GMT
what did the troubleshooting guide tell ya ?
|
|
|
Post by Guest2 on Sept 27, 2022 9:56:02 GMT
what did the troubleshooting guide tell ya ? Nothing about this issue that I went through.
|
|
|
Post by bimitiri on Sept 27, 2022 11:33:32 GMT
Did you check the triac signal is described in the service manual before calibrating the heater temp control ?
|
|
|
Post by Guest2 on Sept 27, 2022 11:47:38 GMT
Yes.
|
|
|
Post by gnurk on Sept 27, 2022 12:03:41 GMT
check that your NTC wires are not damaged due to heat cracked etc, also check that your ntc is telling your machine to send down 120 vac to heater
|
|
|
Post by bimitiri on Sept 27, 2022 13:19:09 GMT
Does the temp matched the 40 degrees dummy connector was plugged onto NTC 3 slot on distribution board ?
Check temp and Atemp reading Debug screen 4 with 40 degrees dummy connector on NTC3 slot on distribution board ?
|
|