I only get a low flow error in dialysis mode but the flow error pops up when you hear a clicking sound on the machine. It’s a 2008T blue star machine.
|
|
|
Post by Sheepdog1 on Sept 29, 2022 21:15:32 GMT
Need more detailed info, how many hours? have you checked the obvious brushes / pressures / diasafe ?
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Sept 30, 2022 1:59:08 GMT
I only get a low flow error in dialysis mode but the flow error pops up when you hear a clicking sound on the machine. It’s a 2008T blue star machine. [brput brand new motors on and calibrated flow, dearatiin motor, inlet flow and just changed diasafe filter this month.
|
|
|
Post by gfreely on Sept 30, 2022 11:56:52 GMT
Could the click be a solenoid/valve?
|
|
|
Post by gnurk on Sept 30, 2022 12:15:01 GMT
low flow is your primary side check everything before balancing chambers
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Sept 30, 2022 18:08:27 GMT
FLow errors you should always start with your loading pressure.... good pressure you've just eliminated the primary side.... Bad pressure it is the primary side that is the issue. Start with that and let us know what you find out.... what is your condo and TMP. That will help in diagnosing the issue.
|
|