Post by Letstrythis on Oct 2, 2022 22:27:36 GMT
Trying to calibrate a 2008T conductivity confirmation cell, acts likes it trying to calibrate then eventually says it is in ERROR. Sometimes says "no comm to meter" at beginning but you can just hit confirm again and it will go ahead and try to calibrate. EMR system is communicating with 232 fine and I tried a new one too with no difference. Here is what I have done:
calibrated temp control calibrated conductivity with OLC enabled rebuilt and calibrated acid pump volume. Giving me a very low volume in calibration like 15 instead of 16.2 and 22.4 instead of 23.4, but runs perfect in conductivity in dialysis mode after calibration. checked temp sensors all good insulation. No corrosion on condo cells, condo cells stable during dialysis. reseated all boards. tried both jug and wall concentrates. Meter setup worked on other machines with no issues, meter is fully charged before calibration.
How much is MCDS $$$$ again?, might give this stuff up.lol PLEASE HELP!!!!!
|
|
|
Post by Sheepdog! on Oct 3, 2022 14:41:44 GMT
try having your 90xl charging as you perform the calibration . helped me with a couple machines
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 231
|
Post by stuff on Oct 4, 2022 12:31:06 GMT
Put the machine in a rinse and let it complete. Then do the calibration
|
|
|
Post by gnurk on Oct 5, 2022 10:05:15 GMT
put machine in dialyze let it run til condi is stable then try the calibration
|
|