Post by Bruno on Oct 4, 2022 16:13:14 GMT
I have a T machine with 16K hours alarming Bibag 12V alarm. Any ideas?
|
|
|
Post by gnurk on Oct 5, 2022 10:02:41 GMT
i would call tech services on this as i am not experienced with the bibag could be power logic brd or bibag interface brd
|
|
|
Post by gfreely on Oct 5, 2022 17:34:42 GMT
That is the signal for a self-test between the Act/Test and the Bibag Interface. Your 12V rail is being interrupted. Corrosion on the Bibag bd maybe?
|
|