Post by Belleville on Oct 6, 2022 12:27:32 GMT
I have a T machine just switch over todo Independent condo test. It pass all test but fail Independent cond. Machine reads 13.7. Calibrated condo cell and pass. The pre Tx conductivity jumps from 10 to 17 and keeps doing it back and forth so its fails test. Replace condo cells same results.
|
|
|
Post by Guest1 on Oct 7, 2022 0:43:15 GMT
Running bi-bag? Bicarb loop? If loop, check bicarb conductivity to make sure it's within range or bring condo closer to 50. You can try the test with a bicarb jug.
Perform acid & bicarb volume cal, then temp & the cond cell calibration with the yellow dongle. That should solve the cond test issue.
|
|
|
Post by BIOMED 304 on Oct 12, 2022 13:25:50 GMT
Sounds weird but check your loading pressure in dialysis mode. I have changed them and it fix that same issue because pressure was running low in dialysis mode.
|
|
|
Post by Justin on Feb 9, 2023 16:18:19 GMT
Having the same issue state above has anyone found a fix for this issue yet
|
|
BMF
Junior Member
Posts: 86
|
Post by BMF on Feb 9, 2023 19:07:08 GMT
Check the insulation on your NTC's as well as the condition of the temperature probes themselves. I have had bad NTC's cause this same condition.
|
|
|
Post by briancanufixthis on Feb 23, 2023 19:54:20 GMT
Did you calibrate with the 90XL and the 232 connector on the back? Didi you try it with the OLC on?
|
|
in the game too long
Guest
|
Post by in the game too long on Feb 28, 2023 1:20:39 GMT
if your on pressure loop it will fail, but if you run just the test again it will pass.
|
|