Post by Teq on Oct 24, 2022 17:54:13 GMT
Folks, I have a K2 machines not turning ON. I already change the two fuses from the power supply board and still doing the same. Any thoughts,, Thanks
|
|
|
Post by gfreely on Oct 24, 2022 18:21:12 GMT
On a K2 I would automatically suspect the Pwr switch (based solely on my previous experiences with K/K2s).
|
|
|
Post by Sheepdog1 on Oct 24, 2022 18:28:19 GMT
power switch or power button on front panel
|
|
|
Post by dave74 on Oct 24, 2022 18:34:31 GMT
Did you check the fuses to see if they were blown? There are some machine problems that could cause your replacement fuses to blow.
|
|
|
Post by Teq on Oct 24, 2022 18:35:56 GMT
On a K2 I would automatically suspect the Pwr switch (based solely on my previous experiences with K/K2s). gfreely, There is any way in particular you will recommend to test the power switch , I will appreciated, Thanks..
|
|
|
Post by Teq on Oct 24, 2022 18:37:45 GMT
Did you check the fuses to see if they were blown? There are some machine problems that could cause your replacement fuses to blow. Yes, I did and one was out, so I replaced both. But the machine still not turning ON.
|
|
|
Post by Teq on Oct 24, 2022 19:12:28 GMT
I fixed, It was the power switch, Thanks to all of you .
|
|
|
Post by gnurk on Oct 25, 2022 13:08:44 GMT
thanks for posting your fix
|
|