Post by GoBlueBaby on Jul 23, 2021 9:40:23 GMT
I have a Bluestar machine with just 2400 hrs. on it it was brought back to technical to have USB replaced being under warranty I had them send me entire display because replacing just that little USB is a real pain. Since replacing display it has NO temp, Ohm' d out heater followed by what TS guide told me to do. Yet I still have NO temperature replaced sensor board pulled out heater bar have 119 volts and yet no temp is it the actuator board? Please help thank you in advance for the recommendations.
|
|
|
Post by gnurk on Jul 23, 2021 12:19:19 GMT
have you tried the NTC 2 which tells the heater to turn on, or the NTC 3 which is for the display also it could be your triac also ohm out your heater rod
|
|
|
Post by gnurk on Jul 23, 2021 12:25:14 GMT
also check the heater switch in back of the machine does it turn off and on crisply or does it seem mushy
|
|
|
Post by dave74 on Jul 23, 2021 13:13:57 GMT
Coincidences do happen, but could anything have happened while you were replacing the display? I can't recall the last time I had a heater totally fail. I guess you checked to be sure you don't have a flow error.
|
|
|
Post by Snollygoster on Jul 23, 2021 13:34:49 GMT
Something from the old forum.
<bubba> posted 14 April 2019 09:54 AM Good evening all-
I have a new bluestar that has 0200 hour on it and it will run for three treatments then the temperature will drop to 27.4 and not come back up. I have re-calibrated temperature and conductivity. Have watch the debug screen and yet to have seen a flow error. I have swamped function boards with a know good board and still have the same issues. I called big blue and they didn't have anything to share other than check you heat rod(which I did before calling them). I will swap the other boards tomorrow. Just wondering if anyone else is experiencing this issue with the Bluestar.. <1987> posted 15 April 2019 08:42 AM If the machine is still under warranty just have field service come out. I have been having all sorts of strange issues with these new machines also. Bopbop91411 posted 15 April 2019 12:46 PM Hide Post I have the same issues with my BS and I just re-calibrate pre-post temp Posts: 130 | Registered: 22 September 2017
|
|
|
Post by GoBlueBaby on Jul 26, 2021 10:10:47 GMT
have you tried the NTC 2 which tells the heater to turn on, or the NTC 3 which is for the display also it could be your triac also ohm out your heater rod I have checked heater switch in back and it was on. I had a FMC tech come out we ohmd out the power logic board of which that was the problem I am now waiting on the board to come in an fix it thank you all for the TS help.
|
|
|
Post by dave74 on Jul 26, 2021 10:54:04 GMT
Thanks for posting your fix.
|
|