Post by Aventura on Dec 1, 2022 14:45:09 GMT
Hi I have a bluestar T machine 1TOS with 5000 hrs giving this message V104 v108 stuck closed after rinse cycle. Replaced bibag hydraulic assy, swapped ALL boards, swapped v24 v26 V104 v108 v105. It seems that after rinse the timer on zero is blinking and flow pump won't stop turning and the bob in the flow indicator is in UP position which should be in down position when rinse program is complete. During rinse bic pressure on debug 15 is achieving more than 400 mmhg w/c is normal but during the vent period (end of rinse) it would go down to negative 400 mmhg, it should only be less than 90 and positive as observed from other known good machines. Another observation is that when it's on service mode the blue conn (spike) shoots water. Any ideas will be very much appreciated..
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Dec 1, 2022 14:51:34 GMT
Try a new function board.
|
|
|
Post by Aventura on Dec 1, 2022 15:52:34 GMT
I've had swapped ALL boards..i ALL boards..this machine is not 2 yrs old..called tech support 3x no success
|
|
|
Post by Aventura on Dec 1, 2022 16:02:10 GMT
Bstardreamer posted same issue on april 25, 2022 were you able to fix this issue Bstardreamer? Can u pls post ur fix thanks
|
|
|
Post by gfreely on Dec 1, 2022 16:18:44 GMT
Check V101?
|
|
|
Post by Aventura on Dec 1, 2022 17:51:35 GMT
V101 is part of the bibag hydraulic assy which is brand new just installed it...
|
|
BMF
Junior Member
Posts: 86
|
Post by BMF on Dec 5, 2022 14:05:41 GMT
Try doing the regulator pressure calibration. Does it go through or do you get an operator error?
|
|
|
Post by Rocket on Dec 13, 2022 18:03:06 GMT
Not saying this is your issue, but I had a constant v104 with some of my bibag machines (we have 7 of the 23 machines in the clinic). We didn't use the bibag feature, we use a loop. Turned out the bicarb loop had a restriction at the beginning of the loop and the pressure sensors on the bibag machines thought v104 was sticking closed.
I did have some issues with another bibag machine with a weird issue. Turned out the little daughter board behind v104 and v105 had acid corrosion on it.
190860 DISTRIBUTION BOX 2 CIRCUIT BOARD
Good luck!!
|
|
|
Post by Aventura on Dec 14, 2022 1:45:22 GMT
Thanks for the input guys,but I've swapped ALL boards that the machine have that you could think of, even the mother board still no luck...
|
|
|
Post by gnurk on Dec 14, 2022 13:09:51 GMT
query?? when you do a COMPLETE rinse, be sure that you hit the confirm key not the enter key
|
|
|
Post by Aventura on Dec 14, 2022 13:22:15 GMT
After complete full rinse i press confirm all the time...i calibrated regulator pressure but the V104/v108 message re appears afterwards. Here's my observation the flow pump won't stop turning or running after complete rinse.
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Dec 14, 2022 21:44:47 GMT
im just amazed you have a 1TOS with only 5k hours and bibag.
|
|
|
Post by Aventura on Dec 15, 2022 5:43:55 GMT
We just bought these 1T0S bluestar premium bibag machines in april 2021 that's why its got 5k hours only
|
|
|
Post by gnurk on Dec 15, 2022 13:04:59 GMT
perhaps you got an out of box failure with your bibag assembly it happened to me x2 with the front panel
|
|
|
Post by Sum1Sumwhere on Dec 21, 2022 15:20:00 GMT
I have been able to fic this probloem with my machines by replacing pressure transducer 110.
|
|