Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Dec 21, 2022 8:25:05 GMT
I have a 9.12 machine that gives a Low TMP alarm, any suggestions guys?
|
|
|
Post by Chuck Weddle on Dec 21, 2022 9:44:12 GMT
Could be a number of things... - 95% of the time it is treatment related.
- Calibrate arterial and venous pressure.
- PDA/its calibration.
|
|
Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Dec 27, 2022 13:44:02 GMT
Could be a number of things... - 95% of the time it is treatment related.
- Calibrate arterial and venous pressure.
- PDA/its calibration.
Thanks, the problem solved after PDA calibration
|
|
|
Post by Admin on Dec 27, 2022 14:30:02 GMT
Thanks for posting your fix!
|
|