Post by guest111 on Dec 22, 2022 19:27:41 GMT
So I have a 2008k that throws a 24v low on startup, swapped out actuator board and the was able to start up the machine. works fine in rinse and heat disinfect, but in dialysis mode no valve cycle and no components will turn on, not sure where to go from here.
|
|
|
Post by guest111 on Dec 22, 2022 19:57:34 GMT
update flow motor and deair motor turn on but no valves cycle swapped out function board with known good board and still no valves cycle.
|
|
|
Post by t1spike on Dec 22, 2022 20:44:05 GMT
check both actuator cable and distribution board
|
|
|
Post by t1spike on Dec 22, 2022 23:44:11 GMT
also check CFS
|
|
|
Post by dave74 on Dec 22, 2022 23:49:19 GMT
Do the acid and bicarb pumps run? If I am reading things right, no water can enter the machine because valve 41 won't open. Even if there were water in the hydrochamber, it couldn't go anywhere because every valve is closed.
Are you certain that the actuator board you put in is good?
|
|