quilla
New Member
Life is Good.
Posts: 25
|
Post by quilla on Dec 26, 2022 18:55:04 GMT
Hello Folks, I have a Fresenius 2008T machine alarming 24V high and Temp over 95 degrees . I tried to calibrate the Temp Sensor but it is giving me a OPERATOR ERROR. Any thoughts? I will appreciated,
|
|
|
Post by Aventura on Dec 27, 2022 1:39:40 GMT
Swapped or replaced a known good power logic board fixed my problem last month on my T machine bluestar premium.
|
|
|
Post by phil on Jan 3, 2023 16:15:55 GMT
power logic board
|
|
|
Post by jimjamjr on Jan 5, 2023 14:10:43 GMT
you can also try just resetting the boards. pull them off of the motherboard, and put them back in firmly for contact. worked for me a couple times.
|
|
|
Post by guest1234 on Jan 25, 2023 16:47:00 GMT
MOTHERBOARD! just FYI they are cheaper than any board inside the card cage.
|
|
quilla
New Member
Life is Good.
Posts: 25
|
Post by quilla on Jan 25, 2023 19:32:51 GMT
Folks , I reset all the boards from the motherboard and that resolve the problems, Thanks to all of you.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Jan 26, 2023 21:26:11 GMT
I've noticed any time I have seen 24v high and the 95 degrees over alarm it has been my mother board had a batch of machines 3tos in my clinic around 8 that solved the problem every time.
|
|