Post by mike101 on Dec 26, 2022 21:52:24 GMT
doing a OLC self test fails. 24. how would one fix this problem?
|
|
|
Post by gnurk on Dec 27, 2022 13:16:56 GMT
query?? what error code are you getting it will help to point you in the right diresction
|
|
|
Post by ThereAreSomeWhoCallMeTim on Dec 27, 2022 17:59:21 GMT
I would start on pg 428 of the troubleshooting guide where it has you run through the self test. there are about 3 paths that it has you take. re-calibrate the conductivity cells, or start swapping boards. one path is power logic, accuator then function, other is power logic, sensor, sensor cables, function or distribution board. If it cancels the test instead of fails the test then there is a list of codes on page 433 that explains what the codes mean (debug 6). codes 1,2 will run you through the same as above. 3-7 will point you to different parts of the machine besides condo issues.
|
|