Post by joey on Dec 27, 2022 15:27:17 GMT
Good Morning All Hope everyone had a great holiday! I started using the RPC pH test strips and I noticed that every machine is 7.6 What is everyone seeing on theirs? Thanks
|
|
|
Post by Guest2day on Dec 27, 2022 18:14:20 GMT
My machines have been 7.2 & 7.4 for a very long time, two different locations. My belief has been that city water, it's location and source of said water plays a big role in machine pH results. Check your city water & RO out pH levels.
|
|
|
Post by MicroTone on Dec 27, 2022 20:43:06 GMT
Those strips are pretty worthless IMHO
|
|
|
Post by Chuck Weddle on Dec 27, 2022 20:50:34 GMT
Good Morning All Hope everyone had a great holiday! I started using the RPC pH test strips and I noticed that every machine is 7.6 What is everyone seeing on theirs? Thanks We typically see about 7.4. What bicarb conductivity and final conductivity do you run?
|
|
joey
Full Member
Posts: 226
|
Post by joey on Dec 27, 2022 23:05:43 GMT
Good Morning All Hope everyone had a great holiday! I started using the RPC pH test strips and I noticed that every machine is 7.6 What is everyone seeing on theirs? Thanks We typically see about 7.4. What bicarb conductivity and final conductivity do you run? we run at 3.2 bicarb conductivity and total 13.8. how about you?
|
|
|
Post by Chuck Weddle on Dec 28, 2022 9:28:25 GMT
3.1 and 13.6 which gives us a CO2 of 34 and Na of 137.
|
|