Post by Deleted on Jul 28, 2021 19:06:01 GMT
Hi everyone,
I've got an issue with one of my 2008k2 machines. After 9 seconds of rinse I'm getting a flow inlet error and now my machine is airlocked and will not circulate water. I've changed my inlet water regulator, heat exchanger and have changed actuator board and wire. I've made sure that the float switch is working correctly as well. I've been going through the valves and they have all been ohm in the 64 range.
Any ideas?
|
|
|
Post by dave74 on Jul 29, 2021 2:23:58 GMT
I've found a lot of the time the cause of a flow inlet error is a bad deaeration motor and/or pump. Does the deaeration motor feel like it's running strong? Do you know when the brushes were last replaced? What about your deaeration pressure?
The other possibility is an obstruction downstream from the deaeration pump. If there is no where for the water to go, you may get a flow inlet error.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Jul 29, 2021 12:44:39 GMT
The question I have is did your flow inlet error cause your air lock or did the air lock cause your flow inlet error? Can you get it to work in other modes? I've cleared air locks before with a 60cc syringe hooked onto the low pressure side of the deare pump. I had the RO shut off recently while I was changing diasafes and it caused 3 machines to air lock which caused flow inlet errors. Once the air lock was gone everything was fine.
|
|
|
Post by gnurk on Jul 29, 2021 14:02:28 GMT
I have found that a good number of flow inlet errors are in actuality flow outlet errors the machine cant tell one from the other check your drain for obstructions and V30
|
|
|
Post by Snollygoster on Jul 29, 2021 14:58:09 GMT
Here's a post from the old forum.
BiomedBill posted 18 March 2019 02:18 AM I have a machine that has an intermittent flow inlet error. It's a T machine, with 25,000 hours. We've replaced dearreation motor with a new motor and pump head, replaced valve's 29,30,39,41. Replaced loading pressure regulator, float switch and replaced water inlet pressure regulator. It's always flow inlet error. It will run from 0 to 300 hours in between flow inlet errors. Any ideas what else it could be? Posts: 10 | Registered: 24 October 2014 <Mr Fixit> posted 18 March 2019 02:32 AM Maybe CFS or a sticky valve <vatech> posted 20 March 2019 12:22 AM Go ahead and rebuild your bc valves because they are probably due. And keep looking for that flow inlet error fix. <Val> posted 21 March 2019 03:07 AM More and more valve failures occurring around the 20k mark. my money is on balancing chamber valve. <pops> posted 26 March 2019 10:56 AM insure that your drain is not obstructed no flow out = flow inlet error Mad
disclaimer my views and no one else Razzer MarkT posted 27 March 2019 07:01 AM Hide Post As pops mentioned, first thing i look at is if the drain is all the way snapped on the the wall box connector. Our connectors are the shut off type and if they start to disconnect, it will obstruct the drain flow and cause a flow error. Posts: 84 | Location: OH | Registered: 16 August 2013 <migity> posted 27 March 2019 11:22 AM If it’s rinse only check to see in service mode under treat options if dialysate sampling is in yes. If it is turn it to no. Problem will be solved BiomedBill posted 30 April 2019 02:13 AM Hide Post I'm really sorry it's taken me so long to reply. I honestly forgot. It turned out to be a balancing chamber valve. Thanks all for the input. Posts: 10 | Registered: 24 October 2014 <pops> posted 30 April 2019 11:49 PM thanks for posting your fix
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Aug 4, 2021 17:34:25 GMT
So I'm still working on this one, not sure why this is happening but I was checking the float switch and had a tube going from the vent tube into a graduated cylinder. After checking the cylinder was filled with water. I put the machine back into rinse and it was suctioning through the vent tube?? Any ideas on that one?
|
|
|
Post by gnurk on Aug 5, 2021 13:01:59 GMT
sounds like your float switch is sticking down letting too much water into hydroblock and going out vent tube
|
|
|
Post by whitechocolate on Aug 6, 2021 16:06:56 GMT
How does the drain hose connect to the wall box? Mine have the fitting that snaps onto the wall box. Its a quick release style. Sometimes they will get stuck with spent dialysate (the nasty white stuff) and that will cause my flow errors.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Aug 16, 2021 14:41:10 GMT
Try replacing #10 this can also cause intermittent flow error.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Aug 16, 2021 16:39:43 GMT
Ended up being the float switch, thanks everyone
|
|
|
Post by gnurk on Aug 18, 2021 21:59:16 GMT
josh thanks for posting your fix
|
|
|
Post by RHBiomed on Aug 24, 2021 19:54:29 GMT
Check your diasafe or replace diasafe I should say....
|
|