Post by Roger dodger on Jan 3, 2023 20:43:58 GMT
In self test, filling optical detector test note art alarm is the field alarm system.
|
|
|
Post by Roger dodger on Jan 3, 2023 20:46:15 GMT
In self test fail OD, No Art alarm.
|
|
|
Post by dave74 on Jan 3, 2023 21:35:24 GMT
What about the arterial and venous pressures when nothing is attached to the luer lock fittings? If the pressures are not zero, or if someone hooks up the monitor lines, the machine will likely fail the optical detector test.
|
|