Post by Clint on Jan 4, 2023 16:31:10 GMT
2008t fails optical sensor test.i have replaced it and it still fails, any ideas?
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 229
|
Post by stuff on Jan 4, 2023 17:46:05 GMT
Try to calibrate your venous pressure
|
|
|
Post by t1spike on Jan 5, 2023 17:47:11 GMT
make sure your ART pressure is at 0 when open to air, the pressure must be able to reach 0 if it can't it will fail optical test
|
|
TheTDoc
New Member
Gotta fix em all
Posts: 31
|
Post by TheTDoc on Jan 13, 2023 4:08:48 GMT
(I found this out through experimentation so take with a grain of salt). put a blood line in the optical detector and if you see blood still sensed adjust the opt det potentiometer until it goes away, then put a piece of paper in and if you don't see blood still sensed adjust until it does come up
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 148
|
Post by phillipashleyludlow on Jan 20, 2023 15:22:16 GMT
(I found this out through experimentation so take with a grain of salt). put a blood line in the optical detector and if you see blood still sensed adjust the opt det potentiometer until it goes away, then put a piece of paper in and if you don't see blood still sensed adjust until it does come up when i was at the "training facility" in Waltham for machine school ( the depot where they work on the hospital machines they swap out) one of the guys actually suggested a similar thing with a nod and a wink
|
|