Post by Citizen G on Jan 5, 2023 20:21:04 GMT
greeting to all,
here is my issue. 2008t bluestar, 9488hrs. fails cond. soft only.
i have replaced Act/test bd. Power logic bd, power control bd. function bd- all calibration performed. Bibag gen 2 bd. and interface bd attach to act/test bd. all to no avail. any and all suggestion are welcome. all that is left is the sensor bd.
my query is that i assumed that the Test bd tunes and detunes the tests to be perform. machine runs fines other than failing cond. soft
|
|
|
Post by gnurk on Jan 6, 2023 14:26:17 GMT
try your sensor and actuator cables, then in t and c mode your sensor and function board i am having issues similar with me it venous low hard will not go to -80 tried everything expect it to be your motherboard try the other things first
|
|
|
Post by Citizen G on Jan 6, 2023 15:07:32 GMT
thank you, i will swapped cables out and see what result i get
|
|