Post by Unknown21 on Jan 9, 2023 19:07:54 GMT
Hello everyone! I have a machine with Cond Offset failure when machine starts up at 50%. Cond cells have been calibrated, cleared alarm but just got pulled this morning for the same problem. Any thoughts?
|
|
|
Post by t1spike on Jan 9, 2023 20:48:16 GMT
reseat actuator board and sensor board
|
|
|
Post by ThereAreSomeWhoCallMeTim on Jan 10, 2023 21:21:32 GMT
I just had this also this morning. Besides what t1spike said. Check distro board for any loose plugs, check where the power attaches into the distro board for any scorch marks. Check your -12V (if beyond 11-13V check +12V, Recalibrate Condo's without errors. If errors in Condo cal then start swapping boards. Sensor, actuator, EEPROM, then function board. Pushing in all my plugs on the distro board and making sure the boards were seated properly is what got rid of mine. Also make sure your bicarb and acid pumps are plugged in fully, one of mine seemed a little pulled out.
|
|
|
Post by Unknown21 on Jan 10, 2023 21:27:46 GMT
Thanks! Ended up being sensor board.
|
|