Post by Alex on Jan 13, 2023 17:10:12 GMT
Machine is giving me an acid pump alarm. When I checked acid pump was making like a grinding noise. I replaced acid pump with a new one and machine ran good. Machine made it to treatment floor and is back with same acid pump alarm. Any ideas on where to start?
|
|
|
Post by Sheepdog1 on Jan 13, 2023 18:01:50 GMT
check actuator cable
|
|
|
Post by dave74 on Jan 13, 2023 19:43:50 GMT
Or actuator board.
|
|
|
Post by Alex on Jan 13, 2023 23:03:24 GMT
That’s guys! Replacing the Actuator test board was the fix.
|
|
|
Post by dave74 on Jan 14, 2023 0:23:23 GMT
Thank you for posting your fix.
|
|