Post by banztek on Jan 26, 2023 23:29:17 GMT
|
|
|
Post by dave74 on Jan 27, 2023 0:36:40 GMT
Sure sounds like the battery ram/time keeper chip on the function board. It's the raised chip that plugs into the socket.
|
|
|
Post by guest1234 on Feb 3, 2023 20:17:27 GMT
Do yourself a favor and replace the function board. Just replacing your battery ram chip will fix it but it seems that when it fails it's because of another issue related to the function board.
|
|
|
Post by dave74 on Feb 3, 2023 21:40:46 GMT
Do yourself a favor and replace the function board. Just replacing your battery ram chip will fix it but it seems that when it fails it's because of another issue related to the function board. All I can say is that in my experience, the battery ram solved the problem.
|
|
|
Post by banztek on Feb 10, 2023 1:05:47 GMT
I replaced battery RAM. At first after replacing the yellow battery RAM and power up machine some of the errors in the picture are still there. I performed hydraulic and sensors calibration and the errors are gone. This machine is also in forced mandatory rinse every time at start up, it will not go to main screen unless mandatory rinse is completed. Battery RAM solved the issues.
|
|
|
Post by guest1234 on Feb 10, 2023 15:10:32 GMT
Replace your function board! when you power the machine on go to service mode first! Enter all settings for your facility also verify info in dialysis mode.
|
|
|
Post by Angel on Mar 11, 2023 20:39:31 GMT
I have the same problem on lost settings. I've replaced the whole function board, new battery RAM, EEPROM, motherboard and even sensor board. I'm still having the same problem. Any thoughts?
|
|
|
Post by t1spike on Mar 12, 2023 13:24:12 GMT
after replacing ram, you must do mandatory rinse then shut off machine, turn back on bring up into conductivity, do a diasafe test /shut machine down bring back up in service mode /reset time /turn off machine then turn back on should be all clear now.
|
|
|
Post by dave74 on Feb 8, 2024 13:56:17 GMT
I have fixed all issues with this but still getting Lost Dial Settings on the display. Any suggestions? Did you replace the battery ram chip on the functional board? What settings have you lost? Did you lose your concentrates only, and everything else is OK?
|
|
|
Post by Technician#! on Feb 8, 2024 14:12:00 GMT
I did replace battery ram chip. Think I only lost concentrates but now not coming up in conductivity so i am going to recalibrate the conductivity cells.
|
|
|
Post by Technician#! on Feb 8, 2024 14:14:09 GMT
Sometimes shows short power down also
|
|
|
Post by Technician#! on Feb 8, 2024 16:30:51 GMT
Issue resolved thanks for the help!
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Feb 13, 2024 13:21:05 GMT
Issue resolved thanks for the help! so, just to confirm, you basically recal'd everything?
|
|