CAN ANY HELP WITH THIS ISSUE. HAVE A 6T BIBAG MACHINE WITH 20,000+ HOURS THAT IS CALLIING FOR A MANDATORY RINSE WHEN POWERED ON, WITH: LOST In-Line FlagsBRAM LOST POWER UP FLAGSLOST MISC SETTINGSLOST SVS/UF SETTINGSLOST SVS/UF SETTINGSLOST MISC SETTINGSLOST BP READINGSLOST SOME CONCENTRATE INFOLOST BATTERY RAM DATAHAS ANYONE COME ACROSS SUCH AN ISSUE? WHAT WAS THE FIX OR INFORMATION TOWARDS A FIX. THIS MACHINE HAS NOT HAD THE HARDWARE UP GRADE. PLEASE ANY INFORMATION WOULD BE GREATLY APPERCIATED.Attachments:
|
|
dan
New Member
Posts: 16
|
HELP
Feb 10, 2023 17:09:46 GMT
Post by dan on Feb 10, 2023 17:09:46 GMT
FIX WAS THE BATTERY ON THE FUNCTION BOARD...
|
|
|
HELP
Feb 12, 2023 13:35:32 GMT
Post by gnurk on Feb 12, 2023 13:35:32 GMT
thanks for posting your fix
|
|
|
HELP
Apr 23, 2024 10:14:05 GMT
Post by chewy79 on Apr 23, 2024 10:14:05 GMT
I also had that issue and I replaced the battery on the Function board but I'm still getting "Lost SVS/UF Settings". I calibrated the UF pump. Anything else I can do?
|
|
|
HELP
Apr 23, 2024 13:29:28 GMT
Post by dave74 on Apr 23, 2024 13:29:28 GMT
This was suggested in another post.
after replacing ram, you must do mandatory rinse then shut off machine, turn back on bring up into conductivity, do a diasafe test /shut machine down bring back up in service mode /reset time /turn off machine then turn back on should be all clear now.
|
|
|
HELP
Apr 23, 2024 16:29:47 GMT
Post by chewy79 on Apr 23, 2024 16:29:47 GMT
I did all of that. Still came back up, then I called tech support. The end user, nurse, has to go into SVS profile while setting up for the patient and put in the settings. Thank you for responding though. All is good now.
|
|
|
HELP
Apr 23, 2024 17:09:55 GMT
Post by dave74 on Apr 23, 2024 17:09:55 GMT
Thanks for posting your fix.
|
|