Post by Squid on Aug 2, 2021 11:35:16 GMT
Hello, I am getting a BiBag leak alarm on a machine with 12,000hrs. There is no leak visible from the bibag. The machine runs fine on a jug with no other alarms. The bibag door was replaced a few months ago so i would rule that out. If I take the bag off and put the same bag back on the alarm goes away and does not return. Also, the alarm always happens shortly after the start of the treatment. I have performed the regulator pressure calibration and all other pressures seem fine as well. Has anyone ever had a similar problem? Any ideas are welcome. Thank you in advance.
|
|
|
Post by whitechocolate on Aug 3, 2021 18:20:48 GMT
We just started using bibags yesterday and I've had 3 machines with the same alarm. Each time I've cycled the power and that corrected the issue. I'm not sure where to look either..
|
|
|
Post by Stuff on Aug 4, 2021 10:24:10 GMT
Had this alarm last week as well. Changed bibag out for new one and issue has not shown back up. Checked machine for an actual leak and there was none
|
|
|
Post by DIATECH on Aug 19, 2021 22:36:54 GMT
tighten the bibag air separator chamber screws on top or the valve might be bad Replace a new zip ties on the 2 bottom hoses
|
|