Post by texasbiomed on Feb 20, 2023 17:44:14 GMT
How do we change the settings on blood pressure module on a T-Machine with Bi-bag? Machine reads, Low Volume Patient? Y(1) N(2) every time you need to check blood pressure. Only happens on this machine.
|
|
|
Post by centex on Feb 21, 2023 13:23:11 GMT
I just had a machine pulled for not reading blood pressures and it gave me the same Low Volume Patient? message when I pressed stat/deflate key. According to operator's manual (pg 248), pressing the stat key while in "Select Program" screen will display this message. I selected N(2) took a bp, turned machine off and on again several times and did not get the massage again. Not sure why yours is happening everytime...
|
|
|
Post by angryazbiomed on Feb 21, 2023 17:52:36 GMT
I have had this happen before, the way i fixed the issue was going to the Test/Options screen in Dialysis Mode and make sure the Low Volume button isn't selected. Sometimes they accidentally push it while the PCTs are cleaning the screens
|
|
|
Post by texasbiomed on Feb 24, 2023 19:11:51 GMT
Thank you centex, I will definitely be looking into that next time I'm at the clinic.
|
|