Post by rt198p on Feb 21, 2023 20:29:54 GMT
Went to reimage a chairside and this is the error its giving me has anyone else seen this error called both IT and Machine Technical services with not help from either. Sorry for the sideways image Attachments:
|
|
|
Post by Maybe on Feb 22, 2023 13:23:40 GMT
I can't see the error, since I'm a guest. But when I have had strange issues with chairside, I have changed the CDX Compact Flash Card on the UI-MICs board. Part #190731. Generally clears most problems that arise when a basic reimage doesn't work.
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 230
|
Post by stuff on Feb 22, 2023 13:58:05 GMT
That error is saying that your machine cpu is not compatible with the image version your trying to use.
Image version needs an x86-64bit cpu
Your cpu is listed as an I686
This seems to be an issue with the Linux operating system. I did a google search of I686 and your error popped up, only it was for a PC install of Linux.
I would do as Maybe suggested and do a new flash memory card
|
|
|
Post by t1spike on Feb 22, 2023 18:33:56 GMT
if you have gen 1 ui boards you have to use image 6.0.56 you will have to contact IT to get
|
|
|
Post by wnybioman on Jan 16, 2024 20:35:10 GMT
What was the fix to your chairside issue rt198p?
|
|
|
Post by rt198p on Jan 19, 2024 16:37:50 GMT
New Umics board fixed issue downloaded Chairside image 6.0.56 still didnt work just went with a new gen 3 board
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Jan 19, 2024 16:56:57 GMT
New Umics board fixed issue downloaded Chairside image 6.0.56 still didnt work just went with a new gen 3 board so the issue with downloading and installing the iso 6.0.56 is that the image for the 32bit 6.0.56 tries to use 1.4 gb of ram, but only has 1gb of ram on the UIMICS board. the thumbdrives custy service hands out has a tmp partition for the system to use as storage during the unpack for installs. if you work for big F, they sent out an automated letter with drives to clinics that had gen1's listed in CSD Table. *updated ISO Version* OS version 6.0.56 6.0.62 32bit is ONLY for gen 1 boards and needs to be obtained from customer service via their formatted thumbdrive.
gen 2/3's use 6.17.17 64bit chairside image.
|
|