Post by Unknown24 on Feb 23, 2023 1:10:15 GMT
Hello Everyone,
I have a machine failing V32 34V 36V and 38V on Valve test. Swapped ACT board and Test board. Machine is also failing positive pressure test. Any suggestions?
|
|
|
Post by dave74 on Feb 23, 2023 4:36:11 GMT
What was the original problem that caused you to perform the valve leak test?
|
|
|
Post by Unknown21 on Feb 23, 2023 16:53:48 GMT
Positive pressure test failed
|
|
|
Post by just on Feb 24, 2023 0:33:10 GMT
positive psi fail is usually broken spring in uf pump. can affect diasafe filter test.
|
|
|
Post by Unknown24 on Feb 24, 2023 23:46:16 GMT
It ended up being a UF Spring!
|
|
|
Post by dave74 on Feb 24, 2023 23:56:51 GMT
Thanks for posting your fix.
|
|