Post by Biomed on Mar 1, 2023 15:40:02 GMT
keep getting flow inlet error on rinse. Replaced kinked/split line on outside of hydrochamber. Didn't work, replaced float swich in hydrochamber, this did not work as well. Any suggestions?
|
|
|
Post by Sheepdog1 on Mar 1, 2023 20:50:21 GMT
give us some more info. how old, how many hours , did you just PM the machine? is the Dear motor spinning during rinse, does it have the error in other modes ? possibly Actuator board
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Mar 2, 2023 17:46:59 GMT
Start with your loading pressure. is your loading pressure good? or bad? Good pressure the problem is on your secondary side and would have eliminated changing that hose and float switch.... Bad pressure it's on your primary side and What you did would make sense. Flow inlet error is an outlet problem. Do you have flow to drain??
|
|
|
Post by just a BioMed on Mar 2, 2023 18:22:41 GMT
Diasafe first, fresh side balancing chamber membrane second. I'm assuming you have done the trouble shooting guide.
|
|