Post by Deleted on Aug 2, 2021 18:51:24 GMT
OLC cancels, Cant calibrate condo cell's stuck at first point. Changed function board and sensor board. Calibrated Temp sensors and temp control. Rebuilt bicarb pump. Changed post condo cell. Any help on this would be great. Possible leak I cant see?
|
|
|
Post by dave74 on Aug 2, 2021 19:23:08 GMT
Did you check your acid pump? Sometimes a broken spring will prevent the conductivity from going to up to 15.5 but allow the conductivity to be OK in the normal range.
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Aug 2, 2021 19:47:17 GMT
I have not checked that yet. I will do that next
|
|
|
Post by DIATECH on Aug 19, 2021 21:41:05 GMT
What code from 1-6 OLC are you getting? When you calibrated your temp control, what was your conductivity? Was it stable? I would not calibrate temp control if your conductivity is all over the place.
|
|