Post by kclive on Mar 7, 2023 16:11:56 GMT
I have a 2008T machine that the conductivity is stuck at 10.0. I have replaced the acid pump and checked all of my hydraulic pressures. What am I missing?
|
|
|
Post by dave74 on Mar 7, 2023 17:04:48 GMT
The troubleshooting guide has a somewhat lengthy section on this. I guess the quickest thing to check would be to see if the machine is actually drawing up concentrate. I know you replaced the acid pump, but is it drawing up acid?
|
|
|
Post by Taco123 on Mar 9, 2023 13:44:18 GMT
Its airlocked. Heat disinfect or rinse should clear it.
|
|