Post by dave74 on Jan 15, 2024 16:55:46 GMT
Were you able to fix the issue? I have a the same issue, but only when I put the machine in dialysis. Hi Chico, The person who started this thread never posted what solved his problem. If you read the entire thread, you will see posts from biomeds who had similar problems but had different solutions.
|
|
|
Post by chico on Jan 15, 2024 17:15:31 GMT
I noticed that, thanks
|
|
|
Post by Quan. on Jan 16, 2024 19:34:19 GMT
hello ive been reading this looking for a answer to this issue. I've tried just about everything. i get the 104/108 stuck closed error turning on machine. the only way ive been able to clear alarm is by opening the bibag n closing it n placing in rinse.
|
|
|
Post by Biomed111 on Feb 26, 2024 22:25:56 GMT
had this issue on bibag machine with 20k hours. I had a ton of issues but i replaced the bibag hydraulic kit with the tan box and v108 stuck went away but v104 stuck did not. replaced the other kit (black box) as well and the problem went away. somehow v104/v108 error was causing an art bp no comm and 24v low as well.
|
|